NI 9361 counter configuration pull-up resistor

Solved!
Go to solution

I recently acquired a 9361 counter and I am trying to make things work on the NI MAX app before I tackle the block diagram. I don't think I am seeing all the configuration options when I click on the 9361 device on MAX. When I right click I do not see a configuration option. 

 

For example, The guide for 9361 says you can enable a 5V pull-up resistor for open collector outputs. I do not see an option anywhere for open collector outputs or for enabling the pull-up.

 

Am I missing a driver? are there options located somewhere else?

0 Kudos
Message 1 of 2
(1,364 Views)
Solution
Accepted by topic author ArieSafari

You will need to access this property via the DAQmx Channel property node called “CI.Freq.LogicLvlBehavior”

0 Kudos