Agilent E3633A DC Power Supply

Hi All,

Hope someone can offer some advice here. I got a Agilent E3633A DC
Power supply second hand from overseas.


I swapped the line voltage to 230 for my country. The fuse in it
currently is 6.3a 250v (reccomnded for 110v operation) but for NZ a
3.15a 250v fuse is reccomended.


When I power it up I get one beep and the display self tests for a
second before going blank and none of the buttons do anything.


Sounds like a controller issue of some sort?


Yes, I know I need to change the fuse, but with no load attached, I
cant imaging it being able to detect it has the wrong fuse intstalled
and should power on ok?


Anyone have any other ideas?


Cheers


-Al


Message 1 of 3
(3,588 Views)

Did you ever figure this out?

0 Kudos
Message 2 of 3
(2,504 Views)

On some instrument you need to move the fuse over to select the right line voltage.  We had a similar problem and found the fuse was in a little holder marker 110V and if you turned it over there was a 220V slot.  Moving the fuse over and replacing solved the issue.  Not sure about this one though, I'd consult the manual.