I work for a lab instruments company which uses a USB to IEEE converter. Normally in MAX, we see the interface and then an instrument aligned underneath it. In this one customer's case, it does not . The software appears to be working because our software sees and controls it. This happens both for a normal domain user and also an administrator.
Any thoughts on what to check ?
Thanks
John Walsh
01-27-2018 05:38 AM
If you did connected the instrument correctly on GPIB bus, and the address is setted well.
Do you try "Scan for instruments"?
01-29-2018 08:54 AM
Hi
Yes. Scan for Instruments did not update the display as expected.
Thanks
John
01-29-2018 10:09 AM
Disconnect all instruments from the bus then begin attaching one at a time. After a new instrument is attached, try the Scan for Instruments. Some GPIB instruments (failing) can lock the bus causing a POLL function to fail.
01-29-2018 10:15 AM
Hi
Thanks for the reply. I should have included more detail. There is a single Interface (interface 0) with a single device connected directly to the converter. Higher level functions within the software that use the NI driver function as expected - for example, numerical readbacks from the instrument. It just do not show as expected in the MAX application.