Instrument not displaying in MAX

Hi 

 

I work for a lab instruments company which uses a USB to IEEE converter.  Normally in MAX, we see the interface and then an instrument aligned underneath it.  In this one customer's case, it does not .  The software appears to be working because our software sees and controls it. This happens both for a normal domain user and also an administrator.

 

Any thoughts on what to check ?

 

Thanks

 

John Walsh

0 Kudos
Message 1 of 5
(2,771 Views)

If you did connected the instrument correctly on GPIB bus, and the address is setted well.

 

No instruments displayed.jpg

Do you try "Scan for instruments"?

 

0 Kudos
Message 2 of 5
(2,742 Views)

Hi 

 

Yes.  Scan for Instruments did not update the display as expected. 

 

Thanks 

 

John

0 Kudos
Message 3 of 5
(2,731 Views)

Disconnect all instruments from the bus then begin attaching one at a time.  After a new instrument is attached, try the Scan for Instruments.  Some GPIB instruments (failing) can lock the bus causing a POLL function to fail.

Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
0 Kudos
Message 4 of 5
(2,728 Views)

Hi 

 

Thanks for the reply.  I should have included more detail.  There is a single Interface (interface 0) with a single device connected directly to the converter.  Higher level functions within the software that use the NI driver function as expected - for example, numerical readbacks from the instrument.  It just do not show as expected in the MAX application.

 

John

0 Kudos
Message 5 of 5
(2,726 Views)