Instrument not found

Windows XP SP3. and ni488.2 v.2.6 seems does not maintain older Agilent spectrum analizers. ;(
0 Kudos
Message 21 of 26
(1,518 Views)

Hi

after searching a bit more I also found the 2.6 driver.

I don't believe that older agilent instruments  are not supported anymore.

Did you again try with slow bustiming?

I also mostly disable auto polling but there should be no srq in finding an instrument.

 

greetings from the Netherlands
0 Kudos
Message 22 of 26
(1,516 Views)
Of course I've tried to slow bus-timing. And even disabled autopolling. On Friday I'll try to replace USB-HS and analizer and find which of them are broken. At workplace USB-HS functioning with RS ESI26 very well.
0 Kudos
Message 23 of 26
(1,515 Views)

Good luck

You seem right. This looks like a broken device.

greetings from the Netherlands
0 Kudos
Message 24 of 26
(1,513 Views)
Today tried my GPIB-HS at work and found that all works good. Now I'm finding another Agilent analizer to check if the my one is broken.
0 Kudos
Message 25 of 26
(1,476 Views)

Just tried to connect another Agilent 8563ec spectrum analizer. With the old one there were 997 instruments found from 0 to 996 with Primary addresses =0. With the new one 13 devices were found with primary addresses from 18 to 30. I've tried all buss-timing values.

Seems it is a driver error.

0 Kudos
Message 26 of 26