Keithley 2601 & Labview

I have Keithley 2601 and I am using attached example vi from NI. But I am getting error 1073807339, and instrument doesn't go into remote state.

Any help in this regard would be  very much appreciated.
0 Kudos
Message 1 of 3
(3,136 Views)

Hi Amjad_F,

 

It looks to be a timeout error. Did you already check these resolved discussions?

 

http://forums.ni.com/t5/LabVIEW/Keithley-2400-voltage-sweep-1073807339-error-at-VISA-write-GPIB/m-p/...

 

http://forums.ni.com/ni/board/crawl_message?board.id=140&message.id=40032

 Additional the timing troubleshoot could help:

 

http://digital.ni.com/public.nsf/allkb/874B379E24C0A0D686256FCF007A6EA0

 

 

Best regards

 

Best regards

 

0 Kudos
Message 2 of 3
(3,090 Views)

Are you certain that your hardware connection is correct?  Can you communicate with the instrument with NI-Max?

 

Try running this simple vi to communicate with the instrument.LV2016 VISAwrc.pngThe other case '-1' passes the inputs right on through to the outputs.  Enter into the write buffer:

*IDN?\n

 See what response the instrument gives.  The response should begin with 'Keithley..."

Help the Community (and future reviewers) by marking posts as follows:
If it helped - KUDOS
If it answers the issue - SOLUTION
0 Kudos
Message 3 of 3
(3,079 Views)