hi everyone,
I am trying to use Agilent 34970A data acquisition/switch unit with Labview 2009 SP1. I installed all drivers required for labview and then tried to run examples that came with installatin file but the VI showed error coded 1073807339. I tried all examples (6) but all of them came out with the same error. I searched the entire forum about this error but unfortunately I could not find any usefull answer about it. Why did this error occur ? Please help me on this issue.
Thanks in advance.
04-06-2012 10:17 AM
What type of connection do you have? RS-232, GPIB, USB?
04-06-2012 11:07 AM - edited 04-06-2012 11:11 AM
I use RS-232 port of the device via KeySpan USB adapter (RS-232 to USB) (model: USA-19HS).
04-06-2012 11:35 AM
Take a look at this: http://digital.ni.com/public.nsf/allkb/E6DFA2B8D7E99F9886256C14005E82F7?OpenDocument
The problem may be your USB to Serial adapter. They are notorious for not translating signals that might be required (ussually hardware handshaking). Also, the COM port may change if you reboot your PC so check your device manager and verify the port number.
If you have a DE-9 COM port on your PC, try that; if it works you may need to try a different USB-Serial adapter.
04-06-2012 11:47 AM
for sure try a different serial-usb adapter... i've got a bunch of them from various manufacturers that i'll swap out when stuff starts not working right. Some work better on XP than 7, others work better with some hardware than others.
Overall, I've had real good luck with the IOGear GUC232A. SiLabs CP2102 USB-UARTs work great too, but using those involves a bit more than just plugging in a device.
04-06-2012 05:30 PM
Before running out and getting a new adapter, do some basic troubleshooting. The timeout can be caused by a mismatch in com settings (baud rate, parity, etc) or using the wrong type of serial cable (null-modem or straight through). Check that everything is correct