07-09-2010 01:30 PM
Hi Jordy,
It would appear that the signal is now railing at the negative signal input range of just below -70mV. This is the lower limit of the device. Can you try to connect the signal to a handheld dmm to get a rough estimate of the actual signal? It appears that your strain signal is actually below -70mV.
Can you try to switch your signals, this will identify if the strain signal from channel one works on channel 2?
Best,
07-23-2010 04:45 AM
Hi Adam,
I've switched the strain gage from channel 2 (wich worked proberly) to channel 1 and stil got -0,0787 as a result.
Any idea's?
07-26-2010 11:33 AM
Hi Jice,
It still sounds like you are saturating your input line. Can you try to read the voltage on a device with a larger input range such as a handheld DMM to find out what the actual value is?
Best,
08-23-2010 08:20 AM
What exactly do i need to measure? Sorry i'm just an composite engineer with no electrical skills but with programming skills
08-24-2010 10:40 AM
Hi Jordy,
Just connect your AI+/- to a handheld DMM rather than connecting to the AI+/- of your DAQ device. Then you can find the actual voltage your device is outputting if it is out of range of your TC device.
Best,
08-25-2010 04:06 AM
when i connect the DMM to the SG's I get the foollowing results
SG at AI (X)+
Voltage Strain
AI (X)+ 1.25V -0.07837
AI (X)+8 0.86V -0.07837
SG at AI (X)+8
Voltage Strain
AI (X)+ 0.86V -0.07837
AI (X)+8 1.25V +-0
What am I doing wrong here?
08-25-2010 08:22 AM
You have a differential voltage of 0.39 V and a common mode voltage of 1.06 V. You are trying to measure this with a mV range on the DAQ device.
1. Your strain gauge or signal conditioning devices are not connected properly or are defective. I am not familiar with your devices but these values seem unlikely.
2. Your DAQ device needs to be set to a range which is compatible with the output of the signal conditioning device.
Lynn