I need to use both trigger-in and trigger-out events.  It appears there is only one front panel port, i.e. SMB PFI<0>.  If I use that port for the trigger-in, what options are there to get a trigger-out signal out of the box (not pxi bus)?

0 Kudos
Message 1 of 2
(2,640 Views)
Solution
Accepted by topic author bjasper

The PXI 6541 actually has four PFI lines which can be used for importing and exporting triggers - though as you've determined, only PFI 0 has a direct front panel connection.  The others (PFI 1-3) are accessible through the Digital Data & Control port, and thus can be easily accessed when using either the CB-2162 or SMB-2163 connector block (or any similar breakout for the DDC port):

 

http://www.ni.com/pdf/manuals/370369b.pdf

http://www.ni.com/pdf/manuals/373660d.pdf

 

Hope this helps!

 

James M.  |  Applications Engineer  |  National Instruments