Can I calibrate PCIe-7852R and NI USB-7856R OEM with NI Calibration Executive?

 

How?

 

Our National Instruments Calibration Executive Software version is 3.5 and there is not procedures to products PCIe-7852R and NI USB-7856R OEM.

 

0 Kudos
Message 1 of 3
(4,079 Views)
Solution
Accepted by topic author antti_lappalainen

Hi,

 

There are several ways calibrate your instruments. 

 

Firstly, the help file for calibrating R series devices in Calibration Executive can be found by navigating to the Calibration Executive Help > Device Calibration Procedure > Data Acquisition (DAQ) Devices > NI Dynamic Signal Acquisition > NI R Series Calibration Procedure

 

You can also follow the instructions found in the following document.

 

http://www.ni.com/pdf/manuals/372004d.pdf

 

Lastly, you can also calibrate your software for R series devices by using the calibration utility installed with the drivers. Search for "calibrate 78xxr device" and run the executable to calibrate your devices.

 

Thank you.

 

Kind regards,

 

Nigel

 

 

Message 2 of 3