03-01-2011 09:23 AM
Hello All,
I have a pancake load cell Cooper LGP310, a USB-6251 DAQ, and a labview program trying to read the measurement.
The output of the load cell is 3mV/V, 10V excitation, lb range is 5 lbs to 150,000 lbs. I will be recording an impact force.
Ive tried hooking up the signal wires differentially and RSE. I cannot see a change in the sensor output when the load cell is exposed to tension. Though when it is hooked up to a trancell ti-500e it outputs the force just fine.
Is this a conditioning issue? Can the DAQ not see changes this small?
I'm a ME with VERY LITTLE exposure to this type of project. Dont think less of me....
Thoughts? Additional info?
03-01-2011 01:01 PM
Hi mckesea,
How much force are you applying to the load cell when you run your program? With 150000 lb capacity and 10 V excitation, it takes at least 30 lb to produce a voltage change that the USB-6251 can even detect, and that's ignoring accuracy. (The USB-6251's sensitivity in the +/-100 mV voltage range is 6 uV; 6 uV * (150000 lb / 30 mV) = 30 lb.)
Also, the USB-6251 doesn't have excitation outputs, so what type of external 10 V excitation source do you have connected?
Brad
03-01-2011 01:05 PM
Hey Brad,
I pull up to 10,000 lbs with no change.
I have a 9V battery hooked up for excitation.
03-01-2011 06:34 PM
Hi mckesea,
How do you have the signals wired to your USB-6251? (i.e. which terminal is connected to which?)
Brad
03-02-2011 08:08 AM
Differentially
I have the Signal + (white wire) to the AI0
Signal - (green wire) to the AI8
RSE
I have the Signal + (white wire) to the AI0
Signal - (green wire) to the AIGROUND
03-02-2011 04:20 PM
Hey mckesea,
If you are wiring differentially, I would recommend using pull-down resistors to AI GND. Take a look at the DAQ M Series User Manual specifically on page 4-12 (58) fo proper wiring techniques.
Lynn
03-03-2011 08:15 AM
I tried the pull down resistors yesterday. Does it matter what size resistor or just that they are the same?
I've learned we have a voltage amplifier, Cooper DCM 460, around here somewhere I have not been able to find it. I understand the need for this piece though I thought I would see some change in output voltage from the load cell even though it is so small a voltage.
Im afraid when I find the amplifier or get a new one, hook it up, and try to read a signal I will get the same results. Nothing.
03-04-2011 10:26 AM
Hi mckesea,
Let's start from the beginning. Are you looking at these measurements in Measurement & Automation Explorer? How have you configured your task? Please include any screenshots necessary.
Lynn
03-04-2011 10:31 AM
I have not attempted to look at it in Measurement and Automation Explorer. I went straight to a vi. I included this vi in my original post. Are you able to see how I configured my task in this vi or should I take some screen shots?
03-04-2011 10:36 AM
Hey mckesea,
You should always attempt hardware trouble with Measurement & Automation Explorer (Start Menu>>All Programs>>National Instruments>>Measurement & Automation). This will allow us to isolate the hardware from the software. Within this software, I would create a task and continue testing from there.
Lynn