10-13-2017 07:25 AM - edited 10-13-2017 07:28 AM
Hi,
This post could be more DAQ releated, but I have posted it in veristand as I came across the issue whilst using the DAQ channels in veristand.
Currently evaluating NI Veristand 17 with a simple LV model. I have mapped my inport and outport to DAQ AI and AO respectively. When I try to deploy the SDF I get an error. All details are shown below:
DAQ hardware used (PXI with MXI express PC control):
Veristand SDF overview:
Error Log:
• Start Date: 13/10/2017 12:41
• Loading System Definition file: "I deleted file name"
• Preparing to deploy the System Definition to the targets...
• Compiling the System Definition file...
• Starting VeriStand PC Engines...
• Initializing TCP subsystem...
• Starting TCP Loops...
• Connection established with target LocalPC.
• Sending reset command to all targets...
• Preparing to deploy files to the targets...
• Starting deployment group 1...
* * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * *
The VeriStand Gateway encountered an error while deploying the System Definition file.
Details:
Error -89126 occurred at Project Window.lvlib:Project Window.vi >> Project Window.lvlib:Command Loop.vi >> Project Window.lvlib:Connect to System.vi
Possible reason(s):
Trigger line requested could not be reserved because it is already in use.
=========================
NI VeriStand: DAQmx Control Task.vi:3900002<append>
Property: SampClk.OutputTerm
Destination Device: PXI-6225
Task Name: PXI-6225_AI
* * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * * *
• Sending reset command to all targets...
• Shutting down VeriStand PC Engines...
• Stopping TCP loops.
Waiting for TCP loops to shut down...
• TCP loops shut down successfully.
• Unloading System Definition file...
• Connection with target LocalPC has been lost.
More details:
* The file deploys properly when the 'Chassis master HW synchronisation device' is set to none
* PXI_Trig0 is not reserved in MAX
* The same error is seen when I try to export the Dev3/ai/sampleclock to the /Dev3/PXI_Trig0 in LabVIEW
10-16-2017 10:35 AM
Hi,
From your screenshots it looks like trigger 0 is used as a sample clock.
Could you change your LabVIEW example but use trigger one instead of 0 and see whether that works?
Many Thanks,
Riley Ilieva
Applications Engineer