We might end up having to set the range high and settling for less resolution. In fact, that might actually be perfectly acceptable...I'll have to crunch through some more numbers to see for sure.
What I have found is that while operating near 5 K, we would like to have a 50 mK or better temperature resolution, so that means we need very low heat generation in both the strain guages and the RTDs. To complicate the matter further, these resistors will be in an insulating vacuum, which will hinder heat dissipation, increasing the effect we want to avoid, and we will take several measurements a minute for days on end.
That 50 mK temperature resolution would translate to about a 2 Ohm resolution on the RTD resistance, if that were our only source of error. The most obvious other error factor is measuring the resistance of the strain guages themselves (which then have to be calibrated to the temperture--thus the RTDs), so I'm thinking a resolution on the order of 100 mOhm on the RTDs would be necessary?
...And the data sheet for PXI 4071 says that at a range of 100 kOhm, I should be able to get 10 mOhm resolution with 10uA current, so maybe that will work (lower digits resoultion--i.e. 6 1/2 instead of 7 1/2--are highly preferable for speed). I'll have to check the RTDs and SGs to get more details on the self-heating to see if 10uA is OK.
Thanks for that suggestion, although I would still like to have a way to control the current without compromising speed or resolution, if you, Chris, or anyone else has any more ideas?
k thnx