what does error -10801 on the test panels for a PCI-5911 mean?
08-04-2010 12:58 PM
Hi reichenbach,
I believe the error refers to the following:
"An error occurred during the calibration process. Possible reasons for this error include incorrect connection of the stimulus signal, incorrect value of the stimulus signal, or malfunction of your DAQ device."
Were you trying any sort of calibration, self-calibration, or reset of the device, or were you just trying to acquire a quick signal in the test panel within MAX? Does the device work properly when running a simple shipping example? If not, it sounds like there may be a problem with the device and it would likely need repair. Hope this helps,
08-05-2010 01:31 PM
I got this error by simply running test panels in NI Max and then 'querying' to read input data. i have nothing connected to the input. i thought that the unit should be able to run error-free without an input.
08-05-2010 01:31 PM
what is a 'simple shipping example?'
08-05-2010 07:22 PM
What was meant was any of the LabVIEW shipping examples found within LabVIEW on Help>>Find Examples.
02-17-2011 03:25 PM
how does one run the self calibration for the PCI-5911 units. the manual says this: "Self-calibration is done using a software command."
where is the software command? can this command be run in NI-MAX? if not, what software do i need?
02-18-2011 03:52 PM
You should be able do this in MAX. Just find your 5911 under Devices and Interfaces, right click on it and select "self-calibrate".